 Post subject: Clock of QTERM-G75
PostPosted: Sun May 23, 2010 10:54 pm 
Offline

Joined: Fri Sep 18, 2009 10:59 pm
Posts: 19
Location: Mumbai, India.
Hi,
I am using QTERM-G75. I wanted to know the accuracy / drift of the in-built clock present in the same.
My application requires time synchronization for periods upto 6 months. Is it reliable to use in-built clock of QTERM-G75 for such applications? In such a case, by what amount can this clock drift?


Top
 Profile  
 
 Post subject: Re: Clock of QTERM-G75
PostPosted: Mon May 24, 2010 6:26 am 
Offline
QSI Support
QSI Support
User avatar

Joined: Wed Mar 08, 2006 12:25 pm
Posts: 881
Location: Salt Lake City, Utah
I do not know that anyone has formally studied that particular issue. I did ask one of the engineers who helped with the clock circuitry if he had any insight and will let you know his response.

I just grabbed a Qlarity terminal that has been on the shelf for a while (where while is somewhere between 6 and 18 months) and the clock is off by about 6 minutes. I realize that this is probably way too informal for any real use by you.

If your G75 will have network access over those 6 months, it could synchronize its clock via SNTP. We have several G75s installed here a QSI that use SNTP once a day or so and their clocks have remained within a second of each other for several years now.

_________________
Jeremy
http://www.beijerinc.com


Top
 Profile  
 
 Post subject: Re: Clock of QTERM-G75
PostPosted: Thu May 27, 2010 12:39 pm 
Offline

Joined: Fri Mar 10, 2006 9:53 am
Posts: 6
Location: QSI Corporation
The drift is primarily a function of the crystal frequency tolerance. For the crystal used on G75 this is 20 ppm. Assuming 30 day month (2592000 seconds) you would get a maximum 51.8 seconds drift in that time.

This neglects the effects of temperature (which can be very significant; the additional drift can be as much as 50 ppm at -10 C or 60 C) and crystal aging. For first order approximation, a minute (maximum) per month is a good figure assuming the unit is sitting close to room temperature for that period of time.

_________________
Michael K. Elwood

Senior Design Engineer

QSI Corporation

http://www.qsicorp.com


Top
 Profile  
 
Display posts from previous:  Sort by  
Post new topic Reply to topic  [ 3 posts ] 

All times are UTC - 7 hours


You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum
You cannot post attachments in this forum

Search for:
Jump to:  
cron
Powered by phpBB® Forum Software © phpBB Group